a) Heating as measured in a 1×3 pixel ROI located at the focus is plotted as a
function of time (dual-echo pulse sequence from Fig.
1, BW2 =
−BW1, under manually de-adjusted linear shim, 0.169
mT/m offset was added to the vertically-oriented, physical-y
orientation). Dotted and dashed curves represent measurements from echo #1 and
echo #2, respectively, while the solid line shows corrected results obtained
through Eq. 9. The corrected results agreed
very well with reference results, shown with ‘○’ symbols and
obtained using a properly-shimmed product GRE sequence. A scatterplot (b) and a
Bland-Altman plot (c) were used to compare the corrected temperature measurements to the
reference ones. From (c): Slope/intercept = 1.01/0.01 and
R2 = 0.996. From (d): Mean difference = 0.06
°C (solid line), 95% confidence intervals = ±0.36
°C (±1.96 standard deviation, dashed lines).